1.

Regarding CT generations: a. In first-generation scanners, the gantry rotated 360o as well as the single detector b. The second-generation scanner is a rotate–translate scanner c. In second-generation scanners, the detectors can cover the whole cross-section in one radiation d. In third-generation scanners, the patient-to-detector distance is greater than in other generation scanners e. In third-generation scanners, data acquisition is continuous

Answer»

a. False. In first-generation scanners, the X-ray source and single detector both moved across the scanning plane at 180° rotation. 

b. True. 

c. False. In second-generation scanners, there are around 30 detectors, which are not enough for the whole cross-section. 

d. False. In fourth-generation scanners, the patient-to-detector distance is greater than in other generations. 

e. True. In third-generation scanners, data collection is continuous for the full 360°.